Avoximeter Calibration Verification Kit
7 followers
0 Likes
Hi all--for those using the Avox, what do you use for Cal-Ver? I usually order CVC 223 Co-Oximeter Calibration Verification Controls from RNA medical 2x per year. I ordered at the beginning of December and the ship date has now been pushed to March. Is there a good alternative you can recommend that includes tHb and O2 Hb? I don't need COHb or MetHb.
Thanks!
Thanks!
10 Replies
Reply
Subgroup Membership is required to post Replies
Join POCT Listserv now
Suggested Posts
Topic | Replies | Likes | Views | Participants | Last Reply |
---|---|---|---|---|---|
iSTAT in NICU -chem 8, CG4 | 6 | 0 | 271 | ||
Rotem Sigma Validation Help | 1 | 0 | 106 | ||
Hemochron Sig Elite use outside of manufacturer temperature range | 2 | 0 | 176 |
I couldn't find any cal-ver kits on IL's website, but I found Eurotrol has a kit. Does anyone here use the Eurotrol kit?
Any help is much appreciated!
sbeliveau@elliot-hs.org
I was recently TJC inspected, and the inspector pointed out that our 6 month calibration verifications for iSTATs were really not required since the iSTAT is considered manufacturer-calibrated.
The Avoximeter 1000E's manual specifically states under its Calibration chapter that the 1000E is a factory-calibrated instrument. Ours have never been re-calibrated or sent for repair.
So, I'm considering discontinuing 6month cal ver studies on the Avox (we would still perform it as part of new-instrument studies). Heresy, I know.
Curious if anyone sees an obvious problem with this change? I don't see that this would conflict with TJC, TN state, or CLIA/CMS.
breana.birmingham@dignityhealth.org
THANK YOU!!
But alas...The statement about Cal Ver criteria not applying to devices that do not allow user calibration. See CAP standard POC.08300. True that the instrument would definitely fall in this category. I might ponder that a little more!!