Our NICU is requesting an instrument to perform gases down to a pH of 6.5. We are looking for best fit for them. Does anyone have opinions of one over the other? I know EPOC cartridges are room temp. That is a great selling point. Wondered if anyone else has any other information or opinion. Thank you!!

7 Replies

The short answer:
The Epoc is basically the iSTAT version 2.  Both were created/invented by the same person.  The Epoc has some improvements.

The Abbott Cal/Vers don't all reach the AMR of all the analytes and alternate products are required to validate them.  We used  Eurotrol for pCO2 for the low end and pO2 on the high end for this reason.  Both were required to keep our NICU happy with the AMR on the ISTAT.
This isn't necessary on the Epoc for us.

Siemens pH Cal/vers MEAN levels range from 6.6 to 7.8, each lot is slightly different.
Abbott pH Cal/Vers MEAN levels ragne from 6.5 to 7.9, each lot is slightly different. 

Concur with Jeremy - iSTAT pH cal ver goes to 6.5.

Our NICU is also looking at the istat for their very low birth weight babies. Has anyone had issues with sampling errors with either the istat or epoc? Thanks!

Lois, 

Having a large NICU in a system that has used or is using the iSTAT, Epoc and Prime Plus:

The NICU samples do have a higher rate of sample errors vs the adult patient samples which is largely due to the users attempting to draw the very minimum amount needed to run the assay and occasionally they miss judge.  This is more noticeable with staff  that have less experience with the system and improves with time.
The iSTAT doesn't perform the cartrdige Cal/Ver until the specimen has been loaded, if there is a problem the sample is lost.
  • The benefit to this is that the user can immediately load the sample and walk away while the analyzer does its thing.
The Epoc cal/ver's the cartridge prior to sample inoculation, thus saving the sample if the cartridge is bad.
  • The downside of this is that the user has to wait 3 minutes before they can load the sample.... which can be addressed in the workflow: scan badge, scan wristband, load cartridge.... then get sample (7 min window to load sample before cartridge is expired)

Both systems can be improperly loaded by the user and result in the analyzer throwing sample errors.





Thank you that is extremely helpful information!

@ Jeremy- Are you using the Care-fill capillary tubes for the neonate EPOCs? If so, I'm wondering how they have been working out for you.
We have them for our neonates, but it seems like getting a proper sample can be difficult. 

@ Gloria,  Yes, we use Epoc's Care-fill capillary tubes collection devices. I've given your question and email address to our primary RT educator.   She will be able help better than I could on the techniques of using them.  Hope you don't mind.

Reply
Subgroup Membership is required to post Replies
Join POCT Listserv now
Kyle Gray
almost 2 years ago
7
Replies
0
Likes
11
Followers
938
Views
Liked By:
Suggested Posts
TopicRepliesLikesViewsParticipantsLast Reply
Anybody have experience with the Actalyke?
Edith Synnefakis
about 12 hours ago
0061
Edith Synnefakis
about 12 hours ago
Cal/Ver Istat pCO2
Autilia Sisti
about 12 hours ago
1099
Marcelino Gabriel
about 12 hours ago
Does a PharmD qualify you as a Technical Consultant
McKenna Chandler
5 days ago
10315
JoAnn Crain
about 12 hours ago