validation for a new iSTAT handheld

19 followers
0 Likes

I have just had to replace my first iSTAT handheld with one that was not working. What is your process for validation of this a new iSTAT handheld for Act-k and CG8+ testing? Thank you!

4 Replies

Hi Ashlee,

What I do to validate a new iSTAT via CG8 and ACT-k:
3 patients on CG8 and ACT-k against the reference meter
CG8 controls in duplicate for 2 days
CG8 Cal/Ver 1x only
ACT-k controls in duplicate for 2 days


Per Abbott iSTAT procedure manual (page 13 and page 15), all that's needed to validate a replacement handheld is to run the external simulator. We used to do a validation that included Cal-Ver, duplicate runs of QC, and patient correlations but now we only run each level of QC per cartridge and run the external simulator. 

To validate a new iSTAT, we proceed with the following:
Day 1: External Simulator, Calibration Verification (1x), run 3 levels of tricontrols in duplicate.
Day 2: External Simulator, run 3 levels of tricontrols in duplicate, run one patient correlation in duplicate.

Since the assays already had initial validation, we do an abbreviated one for new devices.
- External Simulator
- Calibration Verification 1x on those cartridges that have that.  For ACT-K, we would follow the Heparin Linearity procedure in the Abbott technical bulletin.  We started doing this because CAP says you need to do reportable range on every device for validation and this was the best way to accomplish that.
- QC in duplicate over 2 days
- 5 patient correlations with low, normal and high values to be included

Reply
Subgroup Membership is required to post Replies
Join POCT Listserv now
Ashlee Byquist
over 1 year ago
4
Replies
0
Likes
19
Followers
777
Views
Liked By:
Suggested Posts
TopicRepliesLikesViewsParticipantsLast Reply
Humidity Monitoring
Jennifer Gagne
2 days ago
100429
Kathleen David
about 3 hours ago
DTM IQCP
Amy Mark
2 days ago
20160
Amy Mark
about 15 hours ago
POC ammonia test
Jake Fray
6 days ago
00203
Jake Fray
6 days ago