validation for a new iSTAT handheld
18 followers
0 Likes
I have just had to replace my first iSTAT handheld with one that was not working. What is your process for validation of this a new iSTAT handheld for Act-k and CG8+ testing? Thank you!
4 Replies
Reply
Subgroup Membership is required to post Replies
Join POCT Listserv now
Suggested Posts
Topic | Replies | Likes | Views | Participants | Last Reply |
---|---|---|---|---|---|
EPOC scanner issue post software update | 6 | 0 | 226 | ||
PFA-100 | 0 | 0 | 106 | ||
HMS+ vs. iSTAT Method Comparison | 2 | 2 | 365 |
What I do to validate a new iSTAT via CG8 and ACT-k:
3 patients on CG8 and ACT-k against the reference meter
CG8 controls in duplicate for 2 days
CG8 Cal/Ver 1x only
ACT-k controls in duplicate for 2 days
Day 1: External Simulator, Calibration Verification (1x), run 3 levels of tricontrols in duplicate.
Day 2: External Simulator, run 3 levels of tricontrols in duplicate, run one patient correlation in duplicate.
- External Simulator
- Calibration Verification 1x on those cartridges that have that. For ACT-K, we would follow the Heparin Linearity procedure in the Abbott technical bulletin. We started doing this because CAP says you need to do reportable range on every device for validation and this was the best way to accomplish that.
- QC in duplicate over 2 days
- 5 patient correlations with low, normal and high values to be included