I'm just trying to help out our respiratory department by inquiring about this topic.  They use the GEM 5000 instruments and are doing semi-annual comparisons between them and our lab instrumentation.  We are seeing fairly large differences in the Hemoglobin results between the GEMs and the Sysmex.  GEM 5000 proficiency samples look great for Hemoglobin, as well as the Sysmex proficiency Hemoglobin results, they just don't compare well to each other.  What are other facilities doing about this?  What % difference are you using as acceptable for the Hemoglobin results between lab and blood gas instruments?  Are you using a certain sample for this?  They are using patients with lines and drawing a syringe for their GEM 5000 instrument and then an EDTA tube to run on the Sysmex.  Again, just trying to see what others are doing and forwarding this information along.  Any input is appreciated.  Thanks!

5 Replies

The discrepancy may be due to pre analytical errors. 
Are they flushing the line for any contaminants? 
Are the specimens well mixed when testing? 
Is the GEM calculating or doing a direct measurement for the hemoglobin? GEMs can be setup to calculate or directly measured. How was your's setup?
If the GEM is calculating the hemoglobin while the Sysmex directly measures, you will see discrepancies. 

Remember that you can't do method correlations with a calculation. 

GEM 5000 has a cooximeter in it, so total hemoglobin is directly measured, not calculated.  That said, they should be comparing fairly well.  How much difference are you seeing? Also, is there a low or high bias towards one method or the other?  If you have a bias, you may be able to add an adjustment to the slope of one method to match better.  If no bias, and if the difference is within the expected SD/CV of either machine, you may not have a problem.

James - the difference isn't with all of the comparison samples but for example: 
GEM #1 tHb 12.9 GEM # 2 tHb 12.9 GEM # 3 tHb 13.0 Sysmex Hgb 12.0
GEM #1 tHb 12.8 GEM # 2 tHb 13.3 GEM # 3 tHb 12.9 Sysmex Hgb 10.0
GEM #1 tHb 8.7 GEM # 2 tHb 8.9 GEM # 3 tHb 8.7 Sysmex Hgb 7.3
GEM #1 tHb 11.0 GEM # 2 tHb 11.6 GEM # 3 tHb 10.9 Sysmex Hgb 7.5
GEM #1 tHb 8.10 GEM # 2 tHb 8.50 GEM # 3 tHb 8.50 Sysmex Hgb 6.8

Some are a little concerning in the differences, and others not so much  

yes, samples 2 and 4 in particular.  It does appear that the GEM results are trending higher.  THb values that trend higher in blood gas samples can often be attributed to undermixed samples, where the sampling probe aspirates a hemoconcentrated part of the sample and makes the tHb values trend higher - whereas automated cell counters have an automated mixing step by default.  I would start there.  

We run into this occasionally, especially when the hemoglobin is on the lower end. For correlating our GEMs to our Sysmex XN, we use 7% or 0.5 g/dL for acceptable criteria.

Reply
Subgroup Membership is required to post Replies
Join POCT Listserv now
Ashlee Holland
2 days ago
5
Replies
0
Likes
20
Followers
309
Views
Liked By:
Suggested Posts
TopicRepliesLikesViewsParticipantsLast Reply
Temp and Humidity Monitoring
Kristine Zinn
about 8 hours ago
00118
Kristine Zinn
about 8 hours ago
External iSTAT Liquid QC frequency
Andy Y
about 8 hours ago
10150
charmi sakhvala
about 7 hours ago
Aging i.STAT Devices
Beverly Muth
about 16 hours ago
50302
James Beck
about 10 hours ago