External iSTAT Liquid QC frequency

23 followers
0 Likes

Hi everyone,

I wanted know how everyone's liquid QC is set up for their iSTAT. The cartridges we use here are CG4, CG8, & ACT - we QC every shipment, every lot. Then every lot monthly.
I checked the IFU's and confirmed with our Abbott that there is nothing from them that says we have to QC monthly after our initial QC.
I would like to transition to running liquid controls on each cartridge type, but not every lot we have in the hospital.

Or is everyone here still running liquid QC every month, on every lot when there is nothing from manufacture? (I found an old discussion from 2021)

Thanks,
Andy

9 Replies

Hello Andy, 
You can check with you lab or hospital accreditation requirements when manufacture does not have define criteria for liquid QC.
We are accredited by CAP and we follow below requirements for iSTAT liquid quality control, we perform QC for any new shipment, new lot and Monthly as per our IQCP.
  • COM.30450 New Reagent Lot and Shipment Confirmation of Acceptability - Nonwaived Tests: New reagent lots and shipments are checked against previous reagent lots or with suitable reference material before or concurrently with being placed in service.
We do compare old lot and new lot for it's allowable error.
  • Regarding the iSTAT liquid QC, as per CAP standard "POC.07300 Daily QC - Nonwaived Tests":
    • liquid QC is required to get performed on each day when patient testing is performed unless laboratory have IQCP plan approved by the laboratory director defining the control process, including the frequency and use of external and internal controls.
Hope this helps!!


@charmi sakhvala,

Thanks for info on your process! I have a question about your monthly...are you QCing every lot monthly?

We also run per shipment/cartridge type (regulations). We included in our IQCP monthly refrigerator storage checks. Abbott had something in their manual about running level 1 control to check creatinine/Chem8 and level 3 on glucose cartridges. Although the scenario we operate under does not require this, we incorporated it into our IQCP. Those departments with refrigerators that do not contain these cartridges we take one they do hold and rotate the level of control performed each month. 

We have an IQCP and we are in Maryland and perform QC every week because of a state regulation.  Does anyone from Maryland know if that COMAR state regulation for single use devices is still required?  When I visit the regulation it just says "repealed" not sure what that means.

@Andi Y, we perform monthly quality control all open lot numbers in use.

Our IQCP is written for new lot/current lot QC on shipments or same lot, different shipment before use and then monthly QC on all lots in use. If those exist in the POC Lab supply, then we perform that QC on our loaner instruments so those are ready to go. All our areas that utilize ACT, CG4 or CG8 are required to run monthly QC on all the lots present in their areas.

Andy,
You are correct Abbott only requires monthly liquid QC for waived tests. I think over the years only their Creatinine cartridge is still waived. Years ago, we adjusted our IQCP to not do monthly liquid controls on our non-waived cartridges. We had 6+ years of data that showed we never had a cartridge fail liquid quality control. Our system has been through at least 4 rounds of CAP inspections since the change was made. Make sure you have plenty of data to back up your IQCP change. We run liquid QC with every new lot or shipment. Lot to lot is performed with liquid QC versus a previous lot in our bulk storage. The internal simulator is set to run every 8 hours on all our i-STATs. Our hospital is large and we go through a lot of cartridges, and we rarely have more than one lot number at the end of the month and that lot is QC'd with the lot to lot QC. We do monthly cartridge orders as well. For our Creatinine cartridges we will QC all lot numbers monthly but our ACT, CG8+ or CG4, PT+, and CHEM 8 don't get every lot Qc'd every month. Hope this helps.

Thanks everyone for your input and procedures! I am still green and learning as a POCC. I don't want to do something that is deemed non-compliant. 

@Mike Barrett
You bring up a good point about IQCP's. Is an IQCP required for this instance since there is nothing from the manufacture stating anything about monthly QC? Or is this something that can be just changed in my policy?

Thanks again everyone,
Andy

Andy,
IQCP's are a whole different conversation string we could have, but long story short yes, an IQCP is required for the i-STAT. In your RISK assessment and Quality control sections are where you can or can't justify running monthly controls based on your situation. It is really dependent on each Health System's practices. If we kept larger volumes of cartridges in our bulk storage where we had multiple lots for longer periods of time I don't think we would have made the change even though Abbott doesn't require it. Your IQCP is important to justify your practice to inspectors. Policy change would not be enough, but your policy needs to reflect your IQCP. Very few non waived tests don't require an IQCP.  

Reply
Subgroup Membership is required to post Replies
Join POCT Listserv now
Andy Y
21 days ago
9
Replies
0
Likes
23
Followers
710
Views
Liked By:
Suggested Posts
TopicRepliesLikesViewsParticipantsLast Reply
Blood Gas/Chem POC devices
Kristine Zinn
1 day ago
10385
Jeremy MacDonald
about 23 hours ago
Epic-Beaker Venous Draw Requests for Unsolicited Orders
James Beck
1 day ago
30360
Barbara Edmondson
1 day ago
Blood Gas syringe Heparin concentration
Jessica Jenkins
1 day ago
00226
Jessica Jenkins
1 day ago