i-STAT in NICU?
21 followers
0 Likes
Hi!
Can anyone share their experience running i-STAT in the NICU? Our NICU is pushing for an i-STAT for Chem8 cartridge testing. We understand that the Chem8 cartridge is not approved for capillary samples, but we have clinicians insisting that they know it is being run at other hospitals. We're unsure if this is outdated information, wishful thinking, testing as LDT, or if some hospitals have some workarounds.
Can anyone share their experience running i-STAT in the NICU? Our NICU is pushing for an i-STAT for Chem8 cartridge testing. We understand that the Chem8 cartridge is not approved for capillary samples, but we have clinicians insisting that they know it is being run at other hospitals. We're unsure if this is outdated information, wishful thinking, testing as LDT, or if some hospitals have some workarounds.
6 Replies
Reply
Subgroup Membership is required to post Replies
Join POCT Listserv now
Suggested Posts
| Topic | Replies | Likes | Views | Participants | Last Reply |
|---|---|---|---|---|---|
| Hematocrit comparison acceptability criteria | 0 | 0 | 14 | ||
| i-STAT in NICU? | 6 | 0 | 245 | ||
| NOVA StatStrip Glucometers mis-scanning | 5 | 0 | 262 |
We do use the ISTAT in the NICU, but they do NOT use Chem8 cartridges. Their cartridge of choice is the CG8+, with the occasional CG4 (for lactate testing) thrown in. The FDA has not cleared capillary blood use on the Chem8, potassium, ionized calcium, or lactates as of yet. It does sound like it will be quite a while before it will be possibly available once again. Through DE, we prevent those results from posting into the EMR. Unfortunately, we cannot block them from being performed without compromising the NICUs ability to run capillary blood at all.
Short answer: No.
The Abbott i‑STAT CHEM8+ cartridge is not FDA‑cleared for routine clinical use with capillary whole blood.
What the cleared labeling actually says
Why capillary blood sometimes shows up in discussions
That mention does not represent current clearance and is not reflected in the active IFU.
Current regulatory position (what surveyors care about)
Venous whole blood ✅ Yes FDA‑cleared
Arterial whole blood ✅ Yes FDA‑cleared
Capillary whole blood ❌ No Not included in current IFU
Compliance impact (important for POCT programs)
Recommended documentation language (audit‑safe)
Summary answer (the part you usually need for audits and build decisions):
The Abbott i‑STAT CHEM8+ cartridge is CLIA‑moderate complexity overall, with CLIA‑waived status limited to specific analytes and specimen conditions (venous whole blood only).
CHEM8+ on venous whole blood only, waived analytes | Waived
CHEM8+ on arterial blood | Moderate
Mixed specimen handling | Moderate
Hi Tersa,
If this cartridge is not FDA approved for capillary samples then if you are joint commission or CAP accredited it would make it an LDT, meaning you need to do interference testing, ie full validation not verification etc. The clinicians are unaware of all these nuances. This is your strongest counter point.
Adil
----
Adil I. Khan MSc PhD
Professor of Pathology
Medical Director, Point-of-Care Testing
Medical Director, Clinical Chemistry
Temple University Health System
Dept. of Pathology & Laboratory Medicine
Lewis Katz School of Medicine
Temple University
Philadelphia, PA
Tel: 215-707-0965
Cell: 267-908-2114
________________________________