iSTAT Cartridge lot to lot

26 followers
0 Likes

Hello Everyone!

I wanted to see if anyone will share what y'all are doing with the new cartridge lot to lot.  Now that the iSTATs are changed to an analyzer and QC needs to be performed on each analyzer, I want to confirm my thoughts that we need to do the lot to lot on every iSTAT also?  

Also, with ACT correlations is anyone designating a primary iSTAT and what materials are you using for the ACT correlations?

Thanks so much!

Erin

5 Replies

For istat to istat within the same dept, we are utilizing QC for CG8+ and ACT-C (a 6-month look back). For dept to dept (including main lab), we use one istat per dept and then use our main lab blood gas to complete blood gas patient comparisons (CG8+). 

We only have iSTAT ACT in one department*.

I have been told by Abbott tech support that the i-STAT is NOT an analyzer, despite the info circulating otherwise. Also, my interpretation of COM.30450 is that lot to lot comparison is not required. It reads "New reagent lots and shipments are checked against previous reagent lots or with suitable reference material [QC] before...being placed in service." Emphasis on the "or". I feel it is a tremendous waste of reagents to compare lots on an analyzer that cannot be recalibrated or parameters changed in any way.

I can definitely appreciate that perspective. However, the guidance we've received is that CMS now considers the i-STAT an analyzer rather than a reader, and TJC has communicated that expectation to accredited laboratories. As a result, we're being required to treat it more like our other testing devices, including performing biannual comparison studies.

Although, we are not doing lot to lot. I misread the initial question thinking it was bi-annual comparisons.

Yes, it was TJC the rep mentioned. I'm very grateful to be CAP.

We have a designated reference ISTAT. All new lots ran on that one with the current lot ran on another for check in. We run QC monthly on all the other loaner devices in POC and the areas with instruments run QC monthly. We are CAP.

I use patient samples for ACT patient correlations. You can spike them with heparin if need higher values. 10ul per 0.7 mLs blood from Na Citrate. When ready, add 0.3 mLs CaCl2 to activate, mix 3-5 times and load cartridge. Go up by 10ul increments of heparin in a sample to get higher values. Also, there is a Hemochron control you can use on it to get in the 700-1000 range. It is the ACT+ abnormal. If you crack it and let it sit for 1.5-2 min before loading the cartridge, it will read high, but <1000

Reply
Subgroup Membership is required to post Replies
Join POCT Listserv now
Erin Walters
7 days ago
5
Replies
0
Likes
26
Followers
804
Views
Liked By:
Suggested Posts
TopicRepliesLikesViewsParticipantsLast Reply
How Are You Managing POCT Documentation?
AJ Bautista
about 5 hours ago
90219
Travis Cluckey
about 2 hours ago
Abbott HIV 4th gen kit shortage - alternate testing?
Jake Fray
about 5 hours ago
0089
Jake Fray
about 5 hours ago
WorkDay Learning and QML compatibility
Lila Stevens
about 6 hours ago
11137
Jamie Walden
about 4 hours ago