iSTAT ACT comparisons

1 followers
0 Likes

For those of you that use the i-STAT ACT cartridge and have the HMS Plus on site.  Do you perform 6 month correlations between these instruments?  And if so, what samples do you use for testing?


Thank you,


Jennifer Parsons

6 Replies

We just implemented i-STAT ACT cartridges and previously had the HMS Plus. I was planning on only doing i-STAT to i-STAT correlations and HMS to HMS using QC. Would be curious to see if others are doing i-STAT to HMS.

Our cardiac surgery does ISTAT ACT and Hemochron (I know, it's not the HMS Plus). They very nicely perform both on patients during correlation time. Don't know if that is an option.

I have the perfusionist run comparison samples between I-stat and hms every 6 months aprox. I compare them with did the patient reach the goal anticoagulation state as Positive and under the goal (480 secs) as neg. Then compare them qualitatively.

Hi,


In the past, I would purchase older proficiency tests from API and use for correlations between the Hemochron HMS in the OR and one for the i-STAT in Cath Lab.  My PT was CAP.


Stacie

Thank you!


Jennifer Parsons

I asked CAP about this not long ago.  The standard says if an analyte is performed on more than one platform (regardless of technology), but under the same CLIA they must be correlated every 6 months.


 

Reply
Subgroup Membership is required to post Replies
Join POCT Listserv now
Jennifer Parsons
about 8 years ago
6
Replies
0
Likes
1
Followers
1457
Views
Liked By:
Suggested Posts
TopicRepliesLikesViewsParticipantsLast Reply
Labeling of POCT samples and devices
Felix Ortiz-Plaza
about 13 hours ago
10152
Kimberly Arnesen
about 12 hours ago
LeadCare II sample collection
Leslie Morman
about 15 hours ago
10126
Peggy Lindeman
about 14 hours ago
GH100 Training
Anita Belanger
about 15 hours ago
10150
Kenda Webb
about 14 hours ago