Radiometer ABL 90 Analyzer Acceptable Samples

0 followers
0 Likes

Hello Everyone,


 Our facility is transitioning to ABL90 for Blood Gases, POC Lytes (Na, K), MethHemoglobin & Carbon Monoxide, Lactate and ionized calcium. I was wondering if anyone here are currently using Radiometer ABL 90 analyzers and what other acceptable sample aside from dry heparinized blood gas syringe. I just read that lithium heparinized tube are not acceptable for ionized calcium and currently that's what we are using. Do you guys know of any other alternative? Thank you so much!

3 Replies

Hi Donna,


Here is the info for our Lab SOP. We use serum for ionized calcium on our ABL90:


  Serum from whole blood, collected in SST tube, anaerobically.  Tourniquet must remain on during collection but not longer than 3 minutes. Tube must be filled and not opened until analysis. Centrifuge within 4 hours. Specimen is stable for 70 hours when stored at 40 C.


  Contraindications/Limitations/Sources of variability



  • Electrolyte balanced heparin must be used to prevent the binding of Ca++ ions by the heparin, and an excess of heparin must be avoided.

  • Exposure to air can cause loss of CO2, and raise the pH. Underfilled tubes will also cause this to happen.

  • Do not use liquid heparin as it can dilute the sample and produce reduced results for some parameters, by as much as 10%.

  • Hemolyzed samples (greater than 30 mg/dl) cause clinically significant decreases in Ionized Calcium.

  • Whole blood specimens must be well mixed prior to sampling. Specimens sitting for up to 30 minutes may have settled completely and a thorough mix is required.

We have ABL90s at our hospital.  We only have used syringes.  Your question prompted me to find the answer in the ABL90 FLEX operators manual.  It clearly shows 3 types of samples - safePICO syringe, capillary tube and test tube.  It doesn't however state what type of test tube.  What does the vendor say?  This is a great question.


Kathy


 

Thank you for the feedback. So I have talked to our clinical specialists and he said, lithium heparin tubes are acceptable as long as they are don't have serum separator gel. He said the gel from the tubes may cause problem in the long run. He also advised us to use clot catchers because RT's are using the lithium heparin syringes that contain silicon and silicon can also damage the analyzers when it builds up. 


 


 

Reply
Subgroup Membership is required to post Replies
Join POCT Listserv now
Donna Mob
almost 5 years ago
3
Replies
0
Likes
0
Followers
1018
Views
Liked By:
Suggested Posts
TopicRepliesLikesViewsParticipantsLast Reply
iSTAT in NICU -chem 8, CG4
Kim Ballister
3 days ago
50256
Kim Ballister
2 days ago
Rotem Sigma Validation Help
BRENDA PENA
3 days ago
10103
BRENDA PENA
3 days ago
Hemochron Sig Elite use outside of manufacturer temperature range
Melisa Wagoner
5 days ago
20171
Melisa Wagoner
2 days ago