POC Peeps, I need your expertise. Please bear with me, I am about to sound like the novice that I am. I am having trouble wrapping my mind around the Bi-Annual requirements for the iSTAT .
Background info: We have 9 iSTATs and currently run Chem8 and CG4+ cartridges. I have completed my software updates
Background info: We have 9 iSTATs and currently run Chem8 and CG4+ cartridges. I have completed my software updates
· Method to Method: Study between the iSTAT and the Roche Cobas. I read that a minimum of 5 patient samples can be used to satisfy this.
· Calibration Verification: Using TriControls Cal Ver (L1,3,5) on all 9 istats (both cartridges).
· AMR Study: Satisfied by Cal Ver because it measures low, mid and high.
· Accuracy: Is it satisfied by Cal Ver? It has an established range.
· Precision: Do I need to do a precision run using QC or can I run my 5 patient samples in duplicate when testing method to method?
· iSTAT to iSTAT: I’ve read QC can be used. How much? I cannot find anything that says how much QC to run to satisfy this.
What am I missing? Any tips or tricks you would like to share would be greatly appreciated.
11 Replies
Reply
Subgroup Membership is required to post Replies
Join POCT Listserv now
Suggested Posts
Topic | Replies | Likes | Views | Participants | Last Reply |
---|---|---|---|---|---|
Battery Impacting Abbott Freestyle Glucose Meter Integrity | 0 | 0 | 93 | ||
Training vs. Competency with POC | 3 | 0 | 181 | ||
iSTAT cartridge traceability | 3 | 0 | 221 |
For iSTAT to iSTAT, refer to your institution's policy. If you don't have one, write one. You can use as little as two levels.
· Method to Method: Study between the iSTAT and the Roche Cobas.
1. min of 5 patient correct.
2. You are not testing the ISTATs but the analytes on the cartridge. I just pick two istats each time I do bi-annual. Try to rotate to different devices each time.
· AMR Study: Satisfied by Cal Ver because it measures low, mid and high.
3. Yes
4. New devices not bi-annual.
5. New devices not bi-annual.
· Method to Method: Study between the iSTAT and the Roche Cobas.
1. min of 5 patient correct.
2. You are not testing the ISTATs but the analytes on the cartridge. I just pick two istats each time I do bi-annual. Try to rotate to different devices each time.
· AMR Study: Satisfied by Cal Ver because it measures low, mid and high.
3. Yes
4. New devices not bi-annual.
5. New devices not bi-annual.
· Method to Method: Study between the iSTAT and the Roche Cobas.
1. min of 5 patient correct.
2. You are not testing the ISTATs but the analytes on the cartridge. I just pick two istats each time I do bi-annual. Try to rotate to different devices each time.
· AMR Study: Satisfied by Cal Ver because it measures low, mid and high.
3. Yes
4. New devices not bi-annual.
5. New devices not bi-annual.
· Method to Method: Study between the iSTAT and the Roche Cobas.
1. min of 5 patient correct.
2. You are not testing the ISTATs but the analytes on the cartridge. I just pick two istats each time I do bi-annual. Try to rotate to different devices each time.
· AMR Study: Satisfied by Cal Ver because it measures low, mid and high.
3. Yes
4. New devices not bi-annual.
5. New devices not bi-annual.
· Method to Method: Study between the iSTAT and the Roche Cobas.
1. min of 5 patient correct.
2. You are not testing the ISTATs but the analytes on the cartridge. I just pick two istats each time I do bi-annual. Try to rotate to different devices each time.
· AMR Study: Satisfied by Cal Ver because it measures low, mid and high.
3. Yes
4. New devices not bi-annual.
5. New devices not bi-annual.
For instrument to Instrument there is no required number. Most people do 5 because EP evaluator requires a minimum of 5 points for calculation. I have inspected lab that have done 1 to 5. As long as it is done. As Don said write a policy and follow it.
I usually do the thermal Probe check at this time.
What you have to do bi annually on your i-STAT depends on who inspects you.
CAP require, calibration, AMR, method comparison, and meter to meter. Precision is not required biannually, It is required for new meters or new analyte.
Joint Commission DO NOT require Calibration or linearity biannually. They do not require meter to meter either. They only require comparison of the i-STAT with the main lab ( method to method comparison). They also want to see that you perform thermal probe every 6-months and internal simulation readings.
You seem to be doing what is required biannually, if you are CAP inspected.
I get data from my middleware for Method comparison. Sometimes I find duplicates and I collect those data. Sometimes they perform Troponin and Chem8+ in ER and then turn around and order Troponin and Comp Panel. I try to collect as many as 10 minimum for my comparison. Doing this saves i-STAT cartridges which can be expensive.
'