Good morning!! I have a question about comparability of instruments. If you have a large number of devices (100+) what is your procedure for comparing them?
Thanks so much!
Shea Woodward, MT(AMT)
Point of Care Coordinator
MUSC Health
woodwars@musc.edu
843-792-4007
Thanks so much!
Shea Woodward, MT(AMT)
Point of Care Coordinator
MUSC Health
woodwars@musc.edu
843-792-4007
10 Replies
Reply
Subgroup Membership is required to post Replies
Join POCT Listserv now
Suggested Posts
| Topic | Replies | Likes | Views | Participants | Last Reply |
|---|---|---|---|---|---|
| Does a PharmD qualify you as a Technical Consultant | 0 | 0 | 172 | ||
| Upgrading Middleware Validations Plan (RALS) | 4 | 1 | 357 | ||
| Humidity Monitoring | 10 | 0 | 556 |
If the laboratory uses more than one nonwaived instrument/method to test for a given analyte, the instruments and methods are checked against each other at least twice a year for comparability of results.
Could you please elaborate on how you compare 100 meters twice per year in addition to method comparisons? I cannot seem to wrap my head around what that would look like. Is that all in one facility or multiple facilities?
Shea,
Could you please share what you learn from CAP?
Thank you.
You only have to compare analyzers to each other that all fall under the same CLIA licensure, for most situations this limits you to the analyzers at each facility.
Example:
· Epoc serial number 12345 vs Epoc 56789
After completing this you have now proven that all Epocs are reporting the same results.
Comparing different testing methods; Epoc vs Lab
· Our Lab does not do ABG testing so we are only looking at Chem and H&H.
Thank you for taking the time to share this information. I appreciate your efforts.
I contacted CAP for further clarification, and this was the response.
Thank you for contacting the CAP. We have reached out to CMS regarding this issue; unfortunately, because the Epoc performs nonwaived tests, you must compare all of your instruments. This would not apply if the instruments are in locations that have separate CLIA and CAP numbers, or if the testing was classified as strictly waived.