I'm a new POCC and I'm having to do a corrective action write-up for an API ACT-LR Hemochron survey failure. Would anyone be willing to share what they normally put for a response? No clerical errors, QC passed (no trend present), no specimen handling issues, and no instrument issues. Since the sample could not be repeated to verify, I've had the operator perform QC since the sample types are similar. 

4 Replies

I would call API and asked if they have the same sample leftover. They usually have some leftover samples that you can order with $, if it was not expired. And you can retest the sample. 

Echo what Jenny suggested.  Otherwise, you've covered the obvious areas, clerical, QC, and instrument.  With Hemochron ACT, the nature of how both the QC and PT samples are made and have to be prepared, survey failures are almost always sample-handling related.  Nature of the beast.  If it's only one sample of multiple that failed, and you've shown that the operator can successfully perform QC, it's not a sample handling issue that needs to be addressed with the operator, and you as the POCC have done your due diligence.  Only a very unreasonable inspector would disagree with that.

I agree with those above.

If you could share what value you got in the testing and what was the expected range for the value.  The reason I asked is often with the ACT testing, there is a twostep preparation of the sample.  Often the first step is to hydrate the sample, then dilute it with the Calcium Chloride to have a final solution.  That is how the CAP sample needs to be prepared prior testing.  Hopefully the sample preparation was done in accordance to the instructions.

Based on your information, if QC came in and no clerical errors, that that would rule out application of the sample, device and consumable (pre-analytical, analytical and post-analytical steps).  The next step would be to rule out the sample. 


I think the two-step process would just apply to a tube-type ACT test, where there is a diluent and a CaCl2 solution.  Hemochron LR method would be a cartridge test, where the PT is packaged pretty much like the DirectCheck QC vials.

Reply
Subgroup Membership is required to post Replies
Join POCT Listserv now
Amanda Reeves
over 2 years ago
4
Replies
0
Likes
7
Followers
625
Views
Liked By:
Suggested Posts
TopicRepliesLikesViewsParticipantsLast Reply
Requirements for staff performing training
McKenna Chandler
1 day ago
20259
Michael Bishop
about 14 hours ago
Proper Way to Train
Jo Win
2 days ago
60491
Kim Ballister
1 day ago
switching POC glucose from Roche Inform II to Nova Stat Strip
Rosemary Gray
3 days ago
91508
Audrey Fulcher
1 day ago