ACT-LR Survey Failure - API
7 followers
0 Likes
I'm a new POCC and I'm having to do a corrective action write-up for an API ACT-LR Hemochron survey failure. Would anyone be willing to share what they normally put for a response? No clerical errors, QC passed (no trend present), no specimen handling issues, and no instrument issues. Since the sample could not be repeated to verify, I've had the operator perform QC since the sample types are similar.
4 Replies
Reply
Subgroup Membership is required to post Replies
Join POCT Listserv now
Suggested Posts
| Topic | Replies | Likes | Views | Participants | Last Reply |
|---|---|---|---|---|---|
| Roche Chemstrip specific gravity | 1 | 0 | 72 | ||
| Anybody have experience with the Actalyke? | 0 | 0 | 117 | ||
| Cal/Ver Istat pCO2 | 2 | 0 | 220 |
If you could share what value you got in the testing and what was the expected range for the value. The reason I asked is often with the ACT testing, there is a twostep preparation of the sample. Often the first step is to hydrate the sample, then dilute it with the Calcium Chloride to have a final solution. That is how the CAP sample needs to be prepared prior testing. Hopefully the sample preparation was done in accordance to the instructions.
Based on your information, if QC came in and no clerical errors, that that would rule out application of the sample, device and consumable (pre-analytical, analytical and post-analytical steps). The next step would be to rule out the sample.