CAP standard COM.30450
14 followers
1 Likes
"New reagent lots and shipments are checked against previous lots or with suitable reference material before or concurrently with being placed in service."
How do you satisfy this requirement with the iSTAT instrument and a blood gas cartridge?
We just started using the CG8+ cartridge.
How do you satisfy this requirement with the iSTAT instrument and a blood gas cartridge?
We just started using the CG8+ cartridge.
10 Replies
Reply
Subgroup Membership is required to post Replies
Join POCT Listserv now
Suggested Posts
| Topic | Replies | Likes | Views | Participants | Last Reply |
|---|---|---|---|---|---|
| Roche Chemstrip specific gravity | 0 | 0 | 35 | ||
| Anybody have experience with the Actalyke? | 0 | 0 | 92 | ||
| Cal/Ver Istat pCO2 | 1 | 0 | 181 |
Below is a description of the intent of the question from the checklist.
The purpose of this check is to confirm that the use of new analytic reagent lots and shipments
do not affect patient results. Matrix interferences between different lots of reagents may impact
the calibration status of instruments and consistency of patient results. Improper storage
conditions during shipping of reagents may have a negative impact on their ability to perform or
exhibit the same levels of reactivity as intended.
The minimum extent of the reagent check is described below; however, the check must be at
least as extensive as described in the manufacturer's instructions. The laboratory may determine
the number of specimens tested.
ashlee.byquist@hannibalregional.org
What do you do if the results don't match? Can you calibrate an iSTAT if the values don't match?