iSTAT Capillary BGs on peds patients
14 followers
0 Likes
Wondering if I could possibly pick your brains about capillary blood gases. We have received a few requests over the years between our pediatric OR, nursery, and PICU to perform capillary blood gases, but our pathologists have a few different concerns and I was hoping someone could shed some light to help maybe educate me so I can further discuss it with them. I’d really like to establish capillary blood gases here, but I'm having trouble convincing them to perform a validation.
Their concerns are:
-Accuracy and reliability of results
-Clotted or hemolyzed samples that are undetected by the iSTAT
-Diluted samples from “milking” the tissue
Hoping this group can provide insight into how you've tackled this at your institution, your validation process, and how you’ve handled issues along the way? Thanks :)
Hoping this group can provide insight into how you've tackled this at your institution, your validation process, and how you’ve handled issues along the way? Thanks :)
8 Replies
Reply
Subgroup Membership is required to post Replies
Join POCT Listserv now
Suggested Posts
| Topic | Replies | Likes | Views | Participants | Last Reply |
|---|---|---|---|---|---|
| Does a PharmD qualify you as a Technical Consultant | 0 | 0 | 172 | ||
| Upgrading Middleware Validations Plan (RALS) | 4 | 1 | 357 | ||
| Humidity Monitoring | 10 | 0 | 556 |
Our hospital uses the ISTAT exclusively for blood gas testing. We use venous, arterial and capillary. Of course, each sample type has its own set of normal ranges. We educate proper collection of the samples (avoid air bubbles, allow the sample to flow from the capillary collection tube, not go drip, drip, drip which would introduce ambient air. If sample volume is of concern (as with preemies), capillary collection uses the least amount of blood with minimum waste.
Validation occurred long before I was in point of care so I cannot help there. You cannot compare your results to the other two sources since they do have different ranges though arterial would be the closest. I would not personally volunteer for that. :)
and "Avoid hemolysis (bursting of red cells) due to vigorous massaging or “milking.” Hemolysis will cause an increase in potassium results and a decrease in calcium results."
and "Avoid drawing air into the capillary tube."
and "Test samples collected in capillary tubes immediately to avoid clotting (especially in neonates whose blood may clot more quickly)."
and "Most heparinized capillary tubes are not suitable for electrolyte measurements, especially ionized calcium, due to the high concentration of heparin (50 U/mL or more). Use balanced heparin tubes."
Essentially what the iSTAT system manual is saying is that hemolysis will not have an affect on blood gas values, but if you use a cartridge that also measures potassium then you could have problems. Secondly, clotted specimens on the iSTAT simply error out - there would not be an issue of erroneous results due to clotting, you would instead get <> as a result or "Cartridge Error", etc. The only contaminant for blood gases is exposure to room air and in some instances, ice.
At Children's Hospital Colorado we performed capillary gases extensively, collected via heparinized Saf-Wrap capillary tubes and G3+ cartridges when I was there. On the Sample Type selection screen the testing operator could select "CAP" as the specimen type to apply the appropriate reference ranges. Our reference ranges matched our Arterial ranges. With that said, I have not kept up with the limitations and changes to FDA approval for all cartridge types, and am not able to comment on whether or not the G3+ cartridge is still approved for use with Capillary specimens.
I do have one off site NICU where the RNs only have an iSTAT and have to be trained on capillary collection. This works well here as it is a small group, and they have great educators. One thing that made a HUGE difference for us years ago was switching from using a standard cap tube with open ends to using this Sarstedt product with the iSTAT. It is virtually impossible to get air bubbles in this device, and the sample doesn't run out the other end if tipped too far, etc. It collects the exact volume needed for the cartridge. We have virtually no cartridge errors using this.
Sarstedt Minivette 100uL lithium heparin, capillary collection device