ACT iSTAT testing
23 followers
0 Likes
Good morning,
I used the ACT iSTAT testing for Cath lab and OR and wanted to ask the group what their policy is for validation of new devices and 6 month correlations.
We only have ACT iSTAT in hospital, and do not offer an equivalent ACT in the core lab.
Do you compare all of your iSTATs to each other for 6 month method comparisons? Do you perform any kind of calibration verification or linearity studies? How do you validate a new handheld iSTAT ?
Thanks,
Jake
I used the ACT iSTAT testing for Cath lab and OR and wanted to ask the group what their policy is for validation of new devices and 6 month correlations.
We only have ACT iSTAT in hospital, and do not offer an equivalent ACT in the core lab.
Do you compare all of your iSTATs to each other for 6 month method comparisons? Do you perform any kind of calibration verification or linearity studies? How do you validate a new handheld iSTAT ?
Thanks,
Jake
7 Replies
Reply
Subgroup Membership is required to post Replies
Join POCT Listserv now
Suggested Posts
| Topic | Replies | Likes | Views | Participants | Last Reply |
|---|---|---|---|---|---|
| POC ammonia test | 0 | 0 | 169 | ||
| Glucose Meter Strip Validations | 4 | 0 | 326 | ||
| FOB - moving away from POC | 10 | 0 | 494 |
Coag tests are exempt from ARM verifications.
What is ARM verifications? I've never heard this term. Thanks!
Jake
We do not run Cal Ver material, but Eurotrol just came out with CueSee Coag that could be used.
Replacement i-Stat: run 2 replicates of 2 levels of QC on 2 days (ex. Day 1 = level 1 twice and level 2 twice, Day 2 = level 1 twice and level 2 twice), run 3 patient samples on the new analyzer and an existing analyzer and use the semi-annual method comparison criteria to evaluate.
No calibration verification, AMR verification, or linearity studies.
All analyzers have a manufacturer defined test range (AMR) that they can test for.
Example: a manufacturer of a analyzer that tests for Hemoglobin might *claim* the analyzer can accurately measure sample levels between 0.9 g/dl and 25.0 g/dl.... you must prove this claim by running a linearity that it is accurate between this range... OR you must modify the AMR to something you can prove it is accurate to. I've had to modify AMR ranges of tests on several analyzers because they either are not accurate for the entire range or linearity material isn't available to prove the claimed range.
Coag tests are exempt from validating the AMR every 6 months by CAP.
We follow CAP guidelines and per CAP POC.08600 AMR verification is not required for clot-based coagulation tests, platelet function tests, and
other tests where output is a unit of time or arbitrary reporting unit (rather than measured analyte concentration). For a new device you would need to complete accuracy, precision studies. Depending on your medical director how many samples you need for the method comparison /accuracy (usually 10-20) but up to their discretion.
For new devices, the method is already validated so we do a "mini" validation. We run the external simulator and get the thermal probe reading. I perform a linearity using a Heparin Linearity procedure that Abbott has in their manual. I do this because it is listed in as needed to perform Reportable Range in COM.40300, but I do not do this again every 6 months as it is listed as not required to do an AMR in POC.08600. We then run QC in duplicate over 2 days for Precision. 5 patient correlations are run against the reference ISTAT with the % difference calculated. Our TAE for ACT is 12%.