We are planning on using our Cal/Ver data, which we perform on every i-STAT biannually, and compare them against each other. Then do our method comparisons on a subset of meters. For us this would show the devices compare against each other and that the subset compares against the other methods. How is everyone satisfying this new requirement from JACO?
New Blood Analyzer Requirement from CMS for Laboratories
We are planning on using our Cal/Ver data, which we perform on every i-STAT biannually, and compare them against each other. Then do our method comparisons on a subset of meters. For us this would show the devices compare against each other and that the subset compares against the other methods. How is everyone satisfying this new requirement from JACO?
New Blood Analyzer Requirement from CMS for Laboratories
Joint Commission recently received notification from the US Centers for Medicare & Medicaid Services (CMS) about the classification of the Abbott i-STAT system, which includes the i-STAT 1 blood analyzer device and testing cartridges. Effective immediately, Joint Commission–accredited laboratories must incorporate the i-STAT 1 device into their correlation process every six months, in accordance with Quality System Assessment for Nonwaived Testing (QSA) standard QSA.02.08.01. This includes correlation between all i-STAT 1 devices within each Clinical Laboratory Improvement Amendments certificate.
Historically, only the testing cartridges were subject to correlation every six months to evaluate the results of the same test performed with different methodologies or instruments or at different locations. However, based on the most current information available, the i-STAT 1 device is categorized as a blood analyzer, not a reader.
Joint Commission surveyors will evaluate compliance with this requirement during your next survey.
14 Replies
Reply
Subgroup Membership is required to post Replies
Join POCT Listserv now
Suggested Posts
| Topic | Replies | Likes | Views | Participants | Last Reply |
|---|---|---|---|---|---|
| TEG6s for Pediatrics | 3 | 0 | 157 | ||
| Affinion 2 Dropping Numbers from both Operator ID and Specimen ID When Scanning? | 2 | 0 | 170 | ||
| NovaPlus HCG Pregnancy Test Cassettes - Back order? | 1 | 0 | 173 |
many thanks
We only need to correlate directly measured analytes (we were already doing this), and only to similar sample types (we no longer need to correlate plasma/serum methods to whole blood).
Hopefully the link that I shared opens. What we decided to do based off that link is run all 50 of our istats on the same day using all same lots and preforming QC level 1 &2. For ACT we only used the meters that use ACT cartridges and did the ACT level 1 and 2 QC, again using all same lots of QC and cartridges. I then put them into EP evaluator and used simple precision with the TEA as the cut offs then got them signed off.
I just wanted to circle back to this discussion. In April we found out about this change (our JC inspection was in Feb), then found out CMS was coming the following week for review. We wanted to get ahead of the new interpretation due to CMS inspection, so we gathered all 56 of our istats throughout the week and performed meter to meter testing for chemistry/BG/HH analytes (Chem8/CG8/CG4) to a "reference meter". We used 3 tricontrols and same lot of cartridges. I then compared my reference meter like I have for the past 16 years prior to this change to the BG lab and main labs using 5 patient samples.
It will be time for 6 month correlations again in October and we wanted to revamp if possible. There have been some suggestions made and I am getting caught up on lot numbers (importance).
1. Is it acceptable to only use 3 "samples" (tricontrols) to perform meter to meter testing and then 5 patient samples for the reference meter?
2. Is it acceptable to use historic data even if the reagent and qc lot numbers are not the same?
Thank you for your input and suggestions on how you have addressed this interpretation with working smarter and not harder like we chose to do in April lol...
You can use historical QC results, but I would think you need to use the same lot of QC to mimic using the same patient sample for all iSTATs and run all cartridge types on every iSTAT. Cartridge lot isn't as important in my opinion. However, you can pretty much do what you want with medical director approval. I am unable to do it this way because we have so many iSTATs in areas that use different cartridges. We used the CAL/VER levels and do them on all iSTATs at the time of the CLEW update. If you can't do it all at once, you can do it over the course of 6 months as long as every cartridge type is run on all iSTATs for every sample.
Whatever you choose to do should be approved by your medical director and must be written in your IQCP. Per TJC, the monthly QC does not need to be done on every iSTAT as long as you have the reason spelled out in your IQCP. For example, monthly QC is performed to verify the storage conditions. Therefore, it doesn't need to be done on every iSTAT that shares the cartridges stored there.