istat ACT linearity material

24 followers
0 Likes

We will soon be implementing the i-STAT and will initially be performing only ACT Kaolin and ACT Celite testing. Since the i-STAT ACT methodology is not clot-based, CAP requires us to perform linearity studies at implementation and every six months thereafter.
Does anyone have a recommendation for linearity material they are currently using for i-STAT ACT testing? The only product I have found so far is Eurotrol CueSee Coag. However, the material is unassayed, so we would need to establish our own target values.
I would appreciate any recommendations or feedback on products that have worked well for your laboratory.
Deborah Martuch, MLS(ASCP)
deborah.martuch@hf.org

5 Replies

HI
Attached is the linearity procedure from Abbott.

714547-01E.pdf

Hi Deborah!  When you say linearity studies, are you talking about AMR verification and/or calibration verification?  In POC.08300, there is a note at the very end that says "Single use devices, and other test devices that do not allow user calibration, do not require calibration verification".  The i-Stat does not allow user calibration.  For AMR verification, POC.08600 states "AMR verification is not required for clot-based coagulation tests, platelet function tests, and
other tests where output is a unit of time or arbitrary reporting unit (rather than measured analyte concentration)".  The i-Stat ACT-C is reporting a unit of time (seconds) so I don't think AMR verification is required.  Curious to hear what others have to say!

We perform the Heparin Linearity procedure upon initial validation of a new ISTAT analyzer only. We do not perform that at every 6 months due to POC.08600.

I would concur with Angie.

CAP doesn't require AMR for coag testing based on the guideline. Sorry, I don't have it in front of me but I can follow up if you like. Also, Abbott may have a validation plan to follow and should provide some guidance in that document. I have seen them for blood gases and troponin. They are helpful in providing support. CAP is also very helpful in interpreting their guidelines. Let me know if I can help further.

Reply
Subgroup Membership is required to post Replies
Join POCT Listserv now
5
Replies
0
Likes
24
Followers
298
Views
Liked By:
Suggested Posts
TopicRepliesLikesViewsParticipantsLast Reply
istat ACT linearity material
Deborah Martuch, MLS(ASCP), CPP
about 5 hours ago
50299
Mandye Canales
28 minutes ago
Advice Needed Regarding GEM 100 Result Documentation
Melva Anderson
about 12 hours ago
60251
Melva Anderson
about 6 hours ago
Venous Blood Gas Worklfow
Sarah Spaulding
1 day ago
51304
Harvey Hveem
about 4 hours ago