We have become a 4 hospital system.  Three of those use i-STAT.  There is some debate over comparing methods.  How are you comparing the i-STAT with your chemistry analyzer twice a year for troponin?


 


 


Dee Dee Mason MT (AMT)


Point of Care Coordinator/Laboratory TeamLeader


Phoebe Putney Health System


417 3RD Ave


Albany, Ga. 31701


229-312-6166


Fax: 229-312-6105


Email:
dmason@ppmh.org


 


cid:image005.png@01D16B29.1F262600


 



Disclaimer:

The HIPAA Final Privacy Rule requires covered entities to safeguard
certain Protected Health Information (PHI) related to a person's
healthcare. Information being faxed to you may include PHI after
appropriate authorization from the patient or under circumstances
that do not require patient authorization. You, the recipient, are
obligated to maintain PHI in a safe and secure manner. You may not
re-disclose without additional patient consent or as required by
law. Unauthorized re-disclosure or failure to safeguard PHI could
subject you to penalties described in federal (HIPAA) and state
law. If you the reader of this message are not the intended
recipient, or the employee or agent responsible to deliver it to
the intended recipient, please notify us immediately and destroy
the related message.

12 Replies

Dee Dee,


Even though our POC CLIA is separate from the Main Lab, we perform 5 patient comparisons to the Main Lab twice per year.  We time these with the CLEW updates.  Staff in the ED perform a Calibration Verification and run 5 samples for Troponin that are also sent to the Main Lab analyzer.  Of course we have a special NO CHARGE account so the patient is not billed twice.


thanks,


Carrie

 Dee Dee


At St Mary's Janesville, We discontinued i-stat troponins due to the confusion between methods and ranges with the providers; but when we did use the i-stat troponins we compared by drawing an extra tube on the same patient and running them as closely together as possible on the chemistry analyzer and the i-stat. I also compared spun and unspun and saw little variation on the i-stat result. I would run comparisons for the troponin plus other cartridges at the same time I did the six month cal-ver.

Dee Dee~


I pull 5 plasma samples and one whole blood/green top that comes to main lab.  I rerun plasma samples for better correlation.  I also perform this around the CLEW updates.  Works well.


Thanks, Amber

Amber, you mentioned that you performed comparisons with clew updates. So it would be ok to use qc material as your sample for istat to istat, correct?

Sent via Groupsite Mobile.

I run my troponin correlations with all other iSTAT using a whole blood green top tube, usually left over from an iCA or blood gas syringe and run on iSTAT then spin and run on the Centaur.  We look for clinical concordance as the iSTAT result reads much lower than the Centaur.  I also pull any ER patients that both the iSTAT and Lab testing was done which helps to get the higher values.   ivy

Is there a regulation for the number of sample that you need to do for every 6-month correlation, or it depends on your institution? We are doing 5, I am just wondering if we can decrease the # of samples?


Do you do 1 istat analyzer per sample?


Or 1 sample on multiple iSTAT analyzers?


Thanks


 


 


Don’t we need 20 points for correlation?


 



Jeremy Williford, PBT (ASCP)                                                          


Point of Care Support Specialist, Clinical Laboratory Services


South Georgia Medical Center | 2501 N Patterson St. | Valdosta, GA 31602


229.259.4804 (office) |
229.249-5461 (fax)


 


SGMC-Master-Logo


 


Confidentiality Notice:  This email message, including
any attachments, is for the sole use of the intended recipient(s) and may contain confidential and privileged information. Any unauthorized review, use, disclosure or distribution is prohibited. If you are not the intended recipient, please contact the sender
by reply email and destroy all copies of the original message.



 




From: Kathleen Suayan via POCT Listserv (Groupsite) [mailto:users+1191554@poct.groupsite.com]

Sent: Friday, February 10, 2017 3:51 PM
To: Williford, Samuel 'Jeremy' <jeremy.williford@SGMC.ORG>
Subject: [POCT Listserv] Re: Troponin Comparsion




 


WARNING: This email originated outside SGMC's email system.

DO NOT CLICK: Links or attachments unless you recognize the sender and know the content is safe.



We do 5 because we calculate the statistics ( SD, CV, Bias, etc)  of the comparison in EP evaluator. 


5 is the minimum number of samples it require to get a meaningful result.



I don’t think the number of comparison for the biannual comparison is specified.


The more the better but 20 Troponin will be a little expensive.


Does Abbott define acceptable criteria for concordance values? 


Thanks,


Jennifer

Agree with Pet that performing 20 patient comparisons biannually would be a bit expensive and that the number of comparisons is not specified in any regulation or checklist item.  I think it would also be excessive.     You have probably already run many (20+) patient comparisons during your initial instrument validation and the biannual comparison is not a revalidation.  Also, you are performing proficiency and maybe run verification samples as well as previously analyzed competency samples on many of those instruments so you probably know that instruments are not comparing favorably. 

All of this "How many....what time frame etc." comes down to what is in your procedure. How many does the procedure require? Have you written it differently for the tests that have a lower volume? (I did but the minimum is always 5.)  Have you tried to track patient tests in the computer with draw times with a main lab test? (we have to do this in some really low volume instances especially when time is short at the end of 6 month timeframe.)


Also, the comparison of ISTAT to ISTAT can be done with QC. The CAP standard would prefer patient samples but when you are in numbers of ISTAT that are above 85+...that is too expensive.


Ensure that you are not comparing them across CLIA numbers-especially it PT is used for any of this. We no longer run PT on ISTAT except for ACT so we use no PT for comparisons.


Deanna Bogner 210-297-9657

Reply
Subgroup Membership is required to post Replies
Join POCT Listserv now
Dee Dee Mason
over 8 years ago
12
Replies
0
Likes
1
Followers
3928
Views
Liked By:
Suggested Posts
TopicRepliesLikesViewsParticipantsLast Reply
Rapid molecular tests for Covid, FluAB, RSV and Strep A
Alexandr Markevich
4 days ago
00229
Alexandr Markevich
Hemosonics Quantra CPT Code
Mary Garcia
4 days ago
4099
Mary Garcia
4 days ago
Nova StatSensor Creatinine Meter
Chelley Schneider
5 days ago
30189
Chelley Schneider
4 days ago