Accruacy and Precision
1 followers
0 Likes
To all,
I was cited by CAP for lack of a full validation for all new ISTAT. I have just under 100 handhelds.
Does anyone have a homegrown spreadsheet for Accuracy and precision as well as the standards that you use for them to pass for each parameter?
We are really trying to keep the cartridges to a minimum as well as the tech time.
Also, for those of you that are TJC, do they require a full validation for the ISTAT or can you use your IQCP to outline what you will do? (this is what I was doing)
I was in a TJC house before this and they did not require a full validation.
Any help and/or pointing me to references would be appreciated.
Deanna Bogner
210-297-9657
6 Replies
Reply
Subgroup Membership is required to post Replies
Join POCT Listserv now
Suggested Posts
Topic | Replies | Likes | Views | Participants | Last Reply |
---|---|---|---|---|---|
iSTAT in NICU -chem 8, CG4 | 3 | 0 | 113 | ||
Rotem Sigma Validation Help | 1 | 0 | 81 | ||
Hemochron Sig Elite use outside of manufacturer temperature range | 1 | 0 | 125 |
Abbott put out a suggestion on a workup plan for new/replacement analyzers a couple of years ago. In a nutshell:
Hope this helps. I tried to attach the actual document that I was given but this system kept fighting me.
Debra,
Could you please send a copy of the actual document (or the reference number for the document) to pointofcare@libertyhospital.org?
Thanks,
Vicki Shanks
We do pretty much what Debra stated except we run hypoxic controls from Eurotrol as well.
What standards were you cited for?
In the Abbott I-STAT technical bulletin the i-STAT technical bulletin 715209-00F, Calibration Verification and the i-STAT System under the section Verifying that results have not been affected by maintenance or repair procedures it states:
Repaired and newly purchased analyzers are received with factory calibration. Again, the Electronic Simulator can better assure that the analyzer's most important function is within factor specifications than calibration or control solutions.
Testing calibration verification samples or comparing patient sample results on new or repaired analyzer with an older analyzer will assess cartridge performance only. Any variations in analyzer performance will not be statistically discernible above the performance of the cartridges. when multiple analyzers are to be used a facility, Abbott Point of Care Inc. recommends at least two analyzers in any performance verification studies so that statistics reflect the system.
I am curious which standard they cited you under? It sounds as if there is confusion between initial validation and new/replacement handhelds. Did you question the citation and if so what did CAP say?
I sent that actual item to CAP on the challenge. What came back was this
"COM 30350 COM 40310 Provide your validation for the new ISTAT A full validation is required for all new instruments and the CAP requirements are in the All Common Checklist."
When I called the CAP to ask what I was supposed to do with this standard, the person told me "I think your inspector was over reaching. I have nothing in my notes about doing this on the ISTAT."
The person that inspected me said that she has been doing a full validation for 10 years. when I asked her how many of them had "flunked" she said "None. We do it only because CAP says we have to." CAP has always been behind in how they deal with the ISTAT. Anyone remember when they did not accept the electronic simulator for QC in the early 2000's?
TJC uses IQCP for validations. (That is what I was doing by following the manufacturers' instructions.) I have had several e-mails on this posts from people who have switched for this reason. They did not have as many handhelds as I do either.
For us, it may cost us $20,000 more per year for absolutely no benefit. Just $20,000 worth of busy work; added tech time and used cartridges. I am going to have 20 more ISTAT in the next few months with techs that do bench work as their primary duties.