i-STAT clews/Jams

6 followers
0 Likes

When you download the new clews  for the i-STAT, do you run a validation to see if the new software is accurate? We just updated the clews and are currently doing a validation. Or do you only do a 6 month validation?

27 Replies

Silka,


Great summary- but actually if you are inspected by CAP- CAP does not consider the PT/INR a clot based test so cal/verf is required.  However Abbott does not have a cal/verf linearity kit for the PT/INR and CAP does not have a linearity survey as well.


so per CAP here are the options:





































































































Per CAP:
MATERIALS SUITABLE FOR CALIBRATION VERIFICATION
Materials for calibration verification must have a matrix appropriate for the clinical specimens assayed by that method and target values appropriate for the measurement system. 
Suitable materials may include, but are not limited to:
1.            Calibrators used to calibrate the analytical system
2.            Materials provided by the analytical measurement system vendor for the purpose of calibration verification
3.            Previously tested unaltered patient/client specimens
4.            Primary or secondary standards or reference materials with matrix characteristics and target values appropriate for the method,
5.                Third party general purpose reference materials that are suitable for verification of calibration following reagent lot changes if the material is listed 
               in the package insert or claimed by the method manufacturer to be commutable with patient specimens for the method.   A commutable reference material
                      is one that gives the same numeric result as would a patient specimen containing the same quantity of analyte in the analyic method under discussion; i.e.
              matrix effects are absent.  Commutability between a reference material and patient specimens can be demonstrated using the protocol in CLSI EP14-A3,
6.          Proficiency testing material or proficiency testing validated material with matrix characteristics and target values appropriate for the method values 
                    appropriate for the method.
In general, routine control materials are not suitable for calibration verification, except in situations where the material is specifically designated by the method 
manufacturer as suitable for verification of the method's calibration process.

Karen, do you have something in writing from CAP that states they do not consider the INR test to be a clot based method? 

We only run ACT on i-stat.  Is there any calibration or verification for this test that anyone knows of?  Also is the external simulator only required with a CLEW update per manufacturer?


Thanks.


I just love it that CAP makes the rules that PTINR needed to be cal verified with WHAT?


No  vendor makes  cal ver material  and how can Proficiency Material verified


    that the calibration is good when in itself is an unassayed sample.


Let us not  JUST go through the motion.  CAP has to rethink all these things they are making us do.


Does it really improve the quality of our work just because


     we can show some document that we did cal ver even though it is flawed.


Yes- I have an email from Joan Kosiek


Dear Karen,
The ISTAT for PT/INR is not considered a clot based test. With that said, since there is no material provided by Abbott for the cal verification, the laboratory must use alternative sources. After the CLEW updates, the laboratory must verify that the calibration is still stable. (Sometimes the updates include linearity information). If the laboratory director approves the process of using patient PT samples to verify the calibration, it would be acceptable.
I hope this information is helpful to you.
Regards,
Joan

Joan K. Kosiek, MS,MT(ASCP)SH,MB
Technical Analyst, CAP Accreditation Programs
College of American Pathologists
325 Waukegan Road, Northfield, IL 60093
jkosiek@cap.org
Tel: 800-323-4040 ext. 7316 Dir: 847-832-7316 Fax: 847-832-8316
Web: cap.org

Pet, I have been a CAP inspector for 12 years and have never in my life heard of CAP stating the iSTAT PT/INR is not a clot based coag method. I want that in writing.  I have never even considered citing someone for that. Also what is at play here is PEER inspectors, I think that is the major issue. 

Karen, thanks for sending that letter from Joan. I am still picking my jaw up from off the floor. We do not use the INR cartridges, but that is still a very surprising stance she is taking.  


As far as calibration verification after the CLEW, iSTAT does not require as long as you have shown (with your initial primary validation) that the software does not affect the ranges. 

I do perform cal/verf after each CLEW Update.  In my 38 years I had not seen issues with the ranges.  However with one of the CLEW updates last year we had issues with the CREA cartridge ranges with creatinine yet the CHEM8 crea range was not affected.  I think we pick up issues with the manufacturer's ranges rather than the iSTAT itself.

Is Joan new at CAP? It seems like something drastic has changed in the past year or two on how they are answering these types of questions.

No she has been there for years.

That's just sad, you would think by now she would have figured out how the iSTAT works.

Abbott i-STAT - Sensor Equivalence Letter October 2022_2pgs.pdf
Hello Everyone, 
Abbott maintains that all i-STAT analyzers that pass the electronic simulator are equivalent. However, what has changed is that as of October 2022 there is no longer a "...manufacturer's claim for equivalence of individual assay-specific sensors". Please refer to the sensor equivalence letter attached.

Displaying items 16-27 of 27 in total
Reply
Subgroup Membership is required to post Replies
Join POCT Listserv now
Christopher Manz
almost 6 years ago
27
Replies
0
Likes
6
Followers
5250
Views
Liked By:
Suggested Posts
TopicRepliesLikesViewsParticipantsLast Reply
iSTAT in NICU -chem 8, CG4
Kim Ballister
about 20 hours ago
40157
Kim Ballister
about 15 hours ago
Rotem Sigma Validation Help
BRENDA PENA
about 22 hours ago
1086
BRENDA PENA
about 16 hours ago
Hemochron Sig Elite use outside of manufacturer temperature range
Melisa Wagoner
3 days ago
10142
Melisa Wagoner
about 24 hours ago