We do not perform ionized calcium testing at our hospital, therefore we have no way to correlate the result. Has anyone else had this issue, if so, how did you handle it.


Thank you!

7 Replies


Correlated with our reference lab



For our ACT’s that we do on our iSTATS, since the lab does not perform Activated Clotting Times, we do an iSTAT to iSTAT comparison. We have 2 machines and we do a patient comparison to each other.


 



Kathy Andrews
MLT, RT


Point of Care Coordinator / Coordinatrice des Analyses Hors Laboratoire


Moncton & Miramichi Area /
Régions de Moncton et Miramichi


Horizon Health Network / Reseau de Sante Horizon


Tel: (506)-857-5315    Fax: (506) 857-5325


kathy.andrews@horizonnb.ca


“Be the change that you wish to see in the world.”


  Mahatma
Gandhi



 




From: Cindy Engbert via POCT Listserv (Groupsite) <users+1160715@poct.groupsite.com>


Sent: March 31, 2020 2:00 PM

To: Andrews, Kathy (LAB) (HorizonNB) <Kathy.Andrews@HorizonNB.ca>

Subject: [POCT Listserv] iCa for i-Stat




 


ATTENTION! External email / courriel externe.



 


------- Horizon Health Network Disclaimer -------



This e-mail communication (including any or all attachments) is intended

only for the use of the person or entity to which it is addressed and may

contain confidential and/or privileged material. If you are not the intended

recipient of this e-mail, any use, review, retransmission, distribution,

dissemination, copying, printing, or other use of, or taking of any action in

reliance upon this e-mail, is strictly prohibited. If you have received this

e-mail in error, please contact the sender and delete the original and any

copy of this e-mail and any printout thereof, immediately. Your

co-operation is appreciated.



Le présent courriel (y compris toute pièce jointe) s'adresse uniquement à

son destinataire, qu'il soit une personne ou un organisme, et pourrait

comporter des renseignements privilégiés ou confidentiels. Si vous n'êtes

pas le destinataire du courriel, il est interdit d'utiliser, de revoir, de

retransmettre, de distribuer, de disséminer, de copier ou d'imprimer ce

courriel, d'agir en vous y fiant ou de vous en servir de toute autre façon.

Si vous avez reçu le présent courriel par erreur, prière de communiquer

avec l'expéditeur et d'éliminer l'original du courriel, ainsi que toute copie

électronique ou imprimée de celui-ci, immédiatement. Nous sommes

reconnaissants de votre collaboration.

I had the exact same issue recently. Needed to revalidate my Chem8+ cartridges. Since no instruments in my system analyses iCa, I have gotten in contact with Mayo for iCa correlations. Due to the COVID19, they have suspended sending correlation specimens out for validations. What you will have to do is collect however many patients you plan on analyzing for your iCa validation (typically 20) and make sure they stay capped and spun down within an hour of the collection time. Specimen requirement for iCa as per Abbott is to maintain anaerobic conditions. So ideally, drawing 2 tubes from the same patient would be ideal if during your analysis on the iStat, you happen to open the tube. Do try to get in contact with your reference lab and work out to see if you can create billing codes for the validation specimens.


Hopefully this helps!


 


Jason

You don't need to "correlate" a test to another testing platform if you don't have another platform that performs that same test within your facility.




  • If iSTAT is the only test system that performs iCa in your facility you don't need to correlate it to another outside system.



You do need to validate the results as accurate and precise.  This can be done with the Abbott Tri-Control Cal/Ver for the iSTAT iCa.


If you have multiple platforms of the same type (multiple iSTATs for example) you do need to perform a "instrument to instrument" comparison.

Jeremy, you beat me to this response. The requirement is to prove accuracy, and that can be done in many different ways. For instance, with the cal ver kit for i-STAT. It is NOT required to compare to a reference method. I have verified this with CMS, CAP, and COLA.

Cindy, We do not compare I-STAT  iCa to reference lab or another method.  We do not have another method in house.


We are now TJC inspected and the inspector told us that there is no need to compare I-STAT to I-STAT  since it is the same method.  I-STATs are readers.  Testin g happens in the cartridge.


WE just do I-STAT to Main Lab comparison.

Thank you all for your responses. I greatly appreciate it.

Reply
Subgroup Membership is required to post Replies
Join POCT Listserv now
Cindy Engbert
over 4 years ago
7
Replies
0
Likes
1
Followers
1039
Views
Liked By:
Suggested Posts
TopicRepliesLikesViewsParticipantsLast Reply
iSTAT in NICU -chem 8, CG4
Kim Ballister
3 days ago
50259
Kim Ballister
2 days ago
Rotem Sigma Validation Help
BRENDA PENA
3 days ago
10104
BRENDA PENA
3 days ago
Hemochron Sig Elite use outside of manufacturer temperature range
Melisa Wagoner
5 days ago
20172
Melisa Wagoner
2 days ago