Piccolo Implementation

4 followers
1 Likes

Would anyone please share their implementation process for the piccolo. 
We will be using waived cartridges only.
precision
accuracy
linearity ................etc

Thank you.

3 Replies

If you get any can you share?  thanks!

Yes, I implemented a Piccolo for my urgent care using the CMP waived cartridges and validated for whole blood green top lithium heparin tubes. I use liquid controls BRT Chemistry Control Liquid assays by Abaxis the Piccolo manufacturer that is part of Abbott. The package insert of the controls says Product Number 100-9010. 
For validation, I ran both normal and abnormal controls each day for 10 days then I did a 100 patient method comparison of all analytes to the Vista chemistry analyzer. I then input these into EP evaluator for linear regression analysis. My medical director requires the 100 patient comparisons, CLIA only requires about 30. 
Let me know if you want copies of the QC package insert, my SOP etc. I'm happy to help. Jacob.Fray@dhha.org 

Hi Jake,
   If the Piccolo is a waived, what implementation studies are required?  Precision, linearity and method comparison?  Can you send me your SOP?  My email is Jstewart3@ dh.org.  Thanks,

Reply
Subgroup Membership is required to post Replies
Join POCT Listserv now
Sylvia Lowery
almost 5 years ago
3
Replies
1
Likes
4
Followers
564
Views
Liked By:
Suggested Posts
TopicRepliesLikesViewsParticipantsLast Reply
TEG6s for Pediatrics
Brittany Behrendt
4 days ago
30252
Kristin Cox
3 days ago
Affinion 2 Dropping Numbers from both Operator ID and Specimen ID When Scanning?
Annie Janssen
4 days ago
20213
Lori Lasbury
4 days ago
NovaPlus HCG Pregnancy Test Cassettes - Back order?
Orlando Dominguez
4 days ago
10203
Mi Hui Scales
4 days ago