i-STAT ACT and PT INR Calibration Verification

9 followers
0 Likes

All,

We are using i-stat for PT and ACT. What are you all doing for calibration verifications? I see technical bulletin for ACT but I am not seeing anything for PT? I think I am missing something and need assistance! Any other methods?

4 Replies

We do not perform per CAP POC.08600  
 AMR verification is not required for clot-based coagulation tests, platelet function tests, and other tests where output is a unit of time or arbitrary reporting unit (rather than measured analyte concentration). 
In a 2021 CAP inspection - inspector thought should do for ACTk using tech bulletin, but CAP was called on that date and verified that this was not a requuriement.  I have done this ACTk with new test implementation in the past but do not every 6 months. 

I use the Medtronic HEPLine kit for the iSTAT ACTk linearity, much easier than the Abbott technical bulletin. The kit comes with 5 vials which contain 1-5 units of heparin; you can run a baseline and then each of the heparinized vials. Just need a willing donor to fill the vials. If you want more information just let me know. 
I would check with your regulatory agency or Abbott about the PT, there may not be any requirements since it is a coagulation test. 
Hope this helps, Kerri

Agree with Patsy - not required for coagulation-based assays per CAP.

Gosh, I think I’m driving myself crazy! I somehow completely skipped over POC.08450 and POC.08600. Thank you all for this slap back into reality! 

Reply
Subgroup Membership is required to post Replies
Join POCT Listserv now
Ren Bage
over 3 years ago
4
Replies
0
Likes
9
Followers
1261
Views
Liked By:
Suggested Posts
TopicRepliesLikesViewsParticipantsLast Reply
Investigation for CAP Survey
Danni Zheng
about 6 hours ago
0081
Danni Zheng
about 6 hours ago
IDNOW and only influenza A reported
Tracy Learned
1 day ago
61252
Tracy Learned
about 11 hours ago
GEM IQM3 pO2
Erin Adkins
8 days ago
30298
Alexandra Adams
8 days ago