Afinion Instrument- Moving Instrument for Health Fair- Run QC?

7 followers
0 Likes

We are having a health fair where hbA1c testing will be performed via our Afinion AS100 and Afinion 2.
I always thought that if an instrument was moved QC should be run on it. However, when I read the manual that was not mentioned. 
If you transport instruments what do you do? 

3 Replies

Since it's waived, you're probably not required to QC after each move.  The case could be made though, since it's not intended to be a portable device.  It's supposed to be "placed on a dry, clean, stable and horizontal surface."

Hello Danielle,
The instrument is waived, however I would recommend running at least the quality control to make sure the instrument functions are not affected by the relocation and physical change in environment.

I agree.  I think the case could be made for it not being a portable device, thus needing to be treated like you're 'installing it' each time, which requires QC.