i-STAT hs-TnI Patient correlations

19 followers
1 Likes

Hello POC community!
For those validating the new high sensitivity TnI test cartridge, what analyzer/s are you using for patient correlations?
We have a Roche Cobas 8000 in main lab, but that is a troponin T test, and the i-STAT cartridge is troponin I.
Abbott is telling us that cannot utilize this ("comparing apples to oranges") and do not need patients - use calibrations and QC to validate.
Thoughts? Comments? Suggestions?
Your assistance is appreciated!

Diane Cauteruccio C(ASCP)CM, CPP
Point of Care Testing Coordinator
Laboratory Administration
South Shore Hospital
55 Fogg Road, South Weymouth MA 02190
 

2 Replies

Hi, We just completed our hs-Trop validation on the i-Stat vs. Cobas T. Our Medical Director approved doing a concordance study between the two methods. Basically, comparing Abnormal hs-i to Abnormal hs-T and Normal to Normal likewise, using M and F cutoffs for each. i-stat will only be used in the ED and a baseline will be sent to the Core lab for T.

Thanks Mary!