GEM 5000/7000 Shipment Verification and Carryover Testing

18 followers
0 Likes

I have worked with other point of care devices and have done shipment verification for new lots.
Currently we use the epoc and for shipment verification we run QCs, linearity material and 5 patient samples on the old and new lot.

I'm less sure of how to do shipment verification for the GEM blood gas analyzers.
Since all the QC are run internally and QC is run in the beginning just as you insert the cartridge, what is mandatary that should be done besides documenting receive date, in use date and documenting that all QC passed?

In my SOP, I wrote this: "For the GEM Premier 7000 system, traditional lot‑to‑lot reagent verification is not performed. Upon installation of a new GEM PAK, the analyzer executes automated internal quality management (iQM3), including calibration verification and continuous QC monitoring. The PAK may be used clinically once iQM3 confirms acceptable performance with no active alerts."
I think this statement should be sufficient for an inspector.

What is everyone's process for shipment verification for the GEM paks for the GEM 5000/7000?
Also does anyone monitor carryover testing? I'm not sure of the snapshot below is sufficient for inspectors from the manufacturer's insert.

Attachment.
image.png

1 Reply

Yes I would agree with that. Also, Process Control Solution B is ran after each sample so I think that factors in as well.
Attachment.
image.png

Reply
Subgroup Membership is required to post Replies
Join POCT Listserv now
Jo Win
about 1 month ago
1
Reply
0
Likes
18
Followers
318
Views
Liked By:
Suggested Posts
TopicRepliesLikesViewsParticipantsLast Reply
Blood Gas/Chem POC devices
Kristine Zinn
2 days ago
10416
Jeremy MacDonald
1 day ago
Epic-Beaker Venous Draw Requests for Unsolicited Orders
James Beck
2 days ago
30374
Barbara Edmondson
1 day ago
Blood Gas syringe Heparin concentration
Jessica Jenkins
2 days ago
00229
Jessica Jenkins
2 days ago