Good Morning,

At our facility we are going to be transitioning to a digital refractometer for our urine specific gravity testing.  I was wondering if anyone currently uses the Atago PAL-10S device and if so, what they are utilizing for daily QC and the annual calibration.  Per package insert it only states that "zero-setting" be performed each day of patient testing and upon changing of batteries.  Currently with our traditional refractometers, we have them run 3 samples for daily QC; a DI sample, a 2.7% saline solution, and a 7.5% saline solution.  

In addition to QC, the calibration report included in packaging is good for 1 year after the manufacturing date.  What are you using for the annual calibration?

Thank you so much,

Matt T

1 Reply

You can continue using your saline solutions for daily QC. We are using Quantimetrix Dropper QC 1 & 2 and DI water for zeroing. 

As for the annual calibration, we are using SVT Specific Gravity Calibrators at 1.005 & 1.030. Cat # 118 and website www.sciteck.org. Phone # 800-749-4537. 

Reply
Subgroup Membership is required to post Replies
Join POCT Listserv now
Matthew Thomas
5 months ago
1
Reply
0
Likes
18
Followers
390
Views
Liked By:
Suggested Posts
TopicRepliesLikesViewsParticipantsLast Reply
Syphilis Health Check Diagnostics Direct
Christy Covington
about 13 hours ago
10175
Dana Neverdousky
about 12 hours ago
Anyone local to western Massachusetts use True Metrix Pro glucose monitor?
Tracy Learned
1 day ago
00181
Tracy Learned
1 day ago
Cincinnati, OH area excess i-STAT cartridges
Addison Barham
1 day ago
30321
Ariel Acaylar
about 10 hours ago