Istat replacement instrument
24 followers
0 Likes
I just want to know what others are doing when they get a replacement instrument.
We have two istat and one of them got replaced due to malfunction.
Our policy does not clearly state what need to be done in case of replacements.
Thinking of running simulator, QC, cal/verification and 5 patient samples on the new and the second istat.
Is there anything else i should add/do?
Thank you.
We have two istat and one of them got replaced due to malfunction.
Our policy does not clearly state what need to be done in case of replacements.
Thinking of running simulator, QC, cal/verification and 5 patient samples on the new and the second istat.
Is there anything else i should add/do?
Thank you.
5 Replies
Reply
Subgroup Membership is required to post Replies
Join POCT Listserv now
Suggested Posts
| Topic | Replies | Likes | Views | Participants | Last Reply |
|---|---|---|---|---|---|
| Blood gas syringe back order - pediatrics/NICU | 0 | 0 | 210 | ||
| Super Users for Moderate Complex Blood Gas Analyzer | 7 | 0 | 563 | ||
| iSCREEN Urine Tox Cups | 1 | 0 | 220 |
Up until recently, we were performing a small precision study and correlation between the new analyzer and our primary iSTAT. However recently the FDA changed the classification of the iSTAT making it an actual Chemistry analyzer. Because of this, we are now evaluating our entire procedure and will be increasing our Correlation, Precision and adding a Linearity study. I would recommend checking with your regulatory agency on what their requirements are. We received a letter from TJC letting us know they will start inspecting the new iSTAT process immediately at our next inspection.
Thank you.