p02 Venous Blood Gas on ABL90

7 followers
0 Likes

Hello everyone 

do anyone experiencing  high P02 when running a venous blood gas.   We believe it to be a collection issue, but the units are fight us telling us it's an instrument issue.   It's not all specimens, but mainly from the ED and ICU.   Any advice from past experiences would be greatly appreciated. 

Thank you 

Elliott

4 Replies

If calibration and QC are working, I do not see how it could be attributed to an instrument issue.  What I have seen is higher than usual venous pO2 values but they are simply reflecting the fact that the patient is receiving supplemental O2 delivery (whether their original venous pO2 indicated a need or not) and the values are simply reflecting that.  As you also stated, it could also indicate a collection issue/ambient air contamination.  Are they collecting in standard blood gas syringes and successfully venting (removing ambient air) from the sample?

I agree with James. I see this regularly when the patient is receiving oxygen.  Can you share an example of a high pO2 you're seeing? 

Good Morning 
Thank you, guys, for the advice recently we took over blood gases in the lab when we changed to epic.   @james our VBG are collected in a BD LiHep Green tube.  Thanks for the tip about receiving the O2.  I'm looking at 1 patient who PO2 was 228, but looking at the chart he was on supplemental O2 an hour before the collection, our nurses have a hard time documenting things in the new system.  So I'm not sure if he was still receiving treatment when the VBG was drawn. I'm giving them a little lead way because we just went live with epic. a few months ago.  

At our facility, we routinely draw venous gases in lithium heparin vacutainers.  When testing, we use a syringe and adapter to remove the sample from the tube as anaerobically as possible (we do not remove the cap).  I started a correlation study to validate lithium heparin line draw syringes to give us an option for lower volume draws as our policy states that the vacutainer needs to be full.  I am seeing a discrepancy in the pO2 between the vacutainer and syringe with the vacutainer always higher and anywhere from 20-100+ higher than the syringe for PO2.  We are using a GEM 5000, drawing a lithium heparin non-gel green top and syringe from the same patient from the same draw using a butterfly and waste tube.  Our GEM has been operating normally and the results from the green top and syringe are each repeatable.  I have also run the same tubes and syringes on the EPOC to validate and they match to the GEM.  The PO2 discrepancy seems to be with the lithium heparin no gel vacutainer.  We are currently investigating this further. 

Reply
Subgroup Membership is required to post Replies
Join POCT Listserv now
Elliott Friday
almost 3 years ago
4
Replies
0
Likes
7
Followers
856
Views
Liked By:
Suggested Posts
TopicRepliesLikesViewsParticipantsLast Reply
Requirements for staff performing training
McKenna Chandler
1 day ago
20261
Michael Bishop
about 18 hours ago
Proper Way to Train
Jo Win
3 days ago
60493
Kim Ballister
1 day ago
switching POC glucose from Roche Inform II to Nova Stat Strip
Rosemary Gray
3 days ago
91510
Audrey Fulcher
1 day ago